| piston
| Specifies the beam piston as 0% to 50% for the master and -50% to 50% for the slave.
|
Perform a beam-balance ( master / slave ) specially designed for calibrating bipolar imaging. Attempting to calibrate bipolar imaging using the standard beam-balance is difficult due to the number of variables involved in the calibration. The solution is to fix one polarity and adjust the other polarity.
|
|
| The monitor command bbs balance master (high|low) <piston> will configure the master profile. The master profile will be reduced by the specified beam piston to provide sufficient beam piston for configuration of the slave profile.
|
|
| The monitor command bbs balance slave (high|low) <piston> will configure the slave profile. The slave profile will be reduced, without modifying the master profile, to configure the specified beam piston.
|
|
| Note that typically the maximum beam piston will be 30%. Thus the master profile should be reduced by 15% and then the slave profile should be varied from 12% to -12%.
|
Thermal 1.0 / 1.7x equivalents:
|
|
|
- The monitor command bbs balance (master|slave) (high|low) <piston> does not have an equivalent.
|
|
| Thermal 2 equivalents:
|
|
|
- The monitor command bbs balance (master|slave) (high|low) <piston> does not have an equivalent.
|
COMMAND: bbs balance
|
|
|
ERRORCODE: [06421] The polarity state is invalid
|
|
|
ERRORCODE: [06422] The beam piston is invalid
|